@cabanesc
Copy link

The Argo 3901931 float was deployed in 2017. It is an Arvor float equipped with an SBE41CP (8497) and a Kistler pressure sensor (4940374).

This float had received a MIN/MAX warning starting around cycle 105 with fresher salinity values (-0.35 psu at depth). After analysis, it appears that there is a problem with the pressure sensor, which reports higher pressure values than the actual values.

image
Temperature (left) and Salinity (right) for float 3901931 in function of pressure. Cycles 99 to 112


image
Theta/S diagram for float 3901931. Cycles 99 to 112


Pressure problem appears clearly on these three plots: note the large temperature shift in the thermocline, the salinity minimum shift from 850db to 1150db, the shift in the theta/S diagram with the profiles acquired after cycle 105 that are no longer parallel to the previous profiles.
Pressure does not seem correctable with a simple pressure offset as the magnitude of the pressure error increased with depth.

Based on email exchanges between K. Martini and R. Cancouet, this is probably related to a defect detected in the Kistler pressure sensors in 2016 : a sudden shift in the pressure span (the calibration slope of pressure). The problem concerns CTDs built between January and July 2016, and this particular CTD was built in May 2016.
The magnitude of the pressure span shift, is 1-30 %, pivoting at 0 pressure, and always one sign – resulting in higher reported pressure than actual pressure. Although the pressure sensors were checked during the routine manufacturing process for the defect, there was a small chance that the check would not catch all the affected sensors. More information can be found here:
http://www.argo.ucsd.edu/Kistler_report_SBE.pdf

Because pressure is not correctable, we put this float into the grey list for PRES, PSAL, et TEMP with a flag 4 from cycle 104.

Have you found similar cases?

@gmaze gmaze added the PRES Pressure related issue label Feb 15, 2020
@gmaze gmaze added the example Show case an example of QC issue label Feb 28, 2020
@RomainCancouet
Copy link
Member

this is a MOCCA float

@BSHbirgit
Copy link
Member

I will look for my files, I think I have seen something similar