Energy calibration for the electron-detector #10

Answered by ozel
SteveF161 asked this question in Q&A
Discussion options

You must be logged in to vote

The shape of one of your spectra looks actually quite ok, ignoring the x-axis labels of course.
But it depends on your modifications of the code if it is a real one. The important parameters for tweaking are:

min_g = -20 # steepness of falling edge slope, almost vertical
min_length = 44  # about 0.9 ms at 48 khz sample rate
max_length = 120 # about 2.5 ms, tolerates some alpha-pileup

These are the default settings for the alpha-spectrometer. The electron-detector has much shorter pulses (25 - 100 us max.) as you can see here:

The length values are in numbers of samples, so those depend on the sample rate.
If your spectra were recorded with those same min/max length settings, they could …

Replies: 13 comments

Comment options

You must be logged in to vote
0 replies
Comment options

You must be logged in to vote
0 replies
Comment options

You must be logged in to vote
0 replies
Comment options

You must be logged in to vote
0 replies
Comment options

You must be logged in to vote
0 replies
Comment options

You must be logged in to vote
0 replies
Comment options

You must be logged in to vote
0 replies
Comment options

You must be logged in to vote
0 replies
Comment options

You must be logged in to vote
0 replies
Comment options

You must be logged in to vote
0 replies
Comment options

You must be logged in to vote
0 replies
Answer selected by ozel
Comment options

You must be logged in to vote
0 replies
Comment options

You must be logged in to vote
0 replies
Sign up for free to join this conversation on GitHub. Already have an account? Sign in to comment
Category
Q&A
Labels
None yet
2 participants
Converted from issue

This discussion was converted from issue #10 on December 09, 2020 22:50.